Medical Equipment
Neither iFixit nor this repair information is endorsed by or affiliated with the medical device manufacturer.
Student-Contributed Wiki
An awesome team of students from our education program made this wiki.
Airway Pressure High
Obstruction leading to constant high airway pressure (>PEEP +15 cmH2O) during: > 2 breaths or 5 seconds, whichever is greater; 15 ±1.5 s if less than 2 breaths are triggered).
Possible Causes
  • Airway pressure exceeds the preset upper-pressure limit
  • Kinked or blocked tubing
  • Mucus or secretion plug in the endotracheal tube or airways
  • Patient coughing or fighting ventilator
  • Inspiratory flow rate too high
  • Improper alarm setting
  • Blocked expiratory filter
Obstruction in the patient circuit
The patient circuit is the plastic tubing that connects the patient to the ventilator and delivers gas to and from the patient. Check to make sure there are no kinks in the hose or that there are no obstructions within the tubing.
Incorrect airway pressure value
Check the ventilator settings to ensure that the airway pressure is set to the intended value and that the high-pressure alarm is not due to a change in settings.
Error in alarm settings’ limits
Check the ventilator alarm settings to ensure that the limits for which the alarm activates are not within the operating pressure of the ventilator.
Contact service technician
Battery Voltage Low
Battery voltage too low. Cannot guarantee continued ventilator system operation.
Disconnected from the main power
Plug the ventilator into the mains power of the building to restore power to the onboard batteries.
Batteries are dead
If this alarm message arises you must replace the batteries.
Battery Replacement Guide
Limited Battery Capacity
Less than 10 minutes left of battery operating time.
Batteries are dead
O₂ Concentration Low
Measured O2 concentration is below the set value by more than 5 vol.% or concentration is below 18 vol.% which is independent of settings; gas delivered in the O2 supply line is not O2; O2 sensor faulty or exhausted; O2 cell uncalibrated; gas module for O2 faulty.
O2 concentration is too low
If this alarm message arises you must replace the O2 supply.
O2 Supply Replacement Guide
If using an O2 cell, perform O2 cell adjustment
To adjust the O2 cell:
  • Tap CALIBRATION & TESTS in the extended menu
  • Tap O2 CELL ADJUSTMENT once, then again
  • Tap Yes to perform the O2 cell adjustment
  • Tap OK
Perform a pre-use check
The pre-use check includes tests and measurements of:
  • internal technical functionality
  • barometric pressure
  • gas supply
  • internal leakage
  • pressure transducers
  • safety valve
  • O2 cell/O2 sensor
  • flow transducers
  • battery switch
  • patient circuit leakage, compliance, and resistance
  • alarm state
  • alarm output connection (option)
Contact service technician.
Leakage Too High
Leakage too high. The mask/prongs may not be adjusted properly for the patient or maybe the wrong size.
The patient circuit is the plastic tubing that connects the patient to the ventilator and delivers gas to and from the patient. Check to make sure there are cracks or holes in the tubing, and that the tubing is securely sealed to the ventilator and mask.
Obstruction in the endotracheal tube (if invasive ventilation)
The endotracheal tube is a thin plastic tube that is placed inside the mouth of the patient to deliver oxygen. Check the tubing to ensure that there are no obstructions, cracks, or leaks.
The patient interface is too loose (if non-invasive ventilation)
Check the patient interface (breathing mask) to ensure that it forms an airtight seal around the patient’s mouth and that it is delivering oxygen to the patient.
Expiratory Cassette Error
Technical problem with the expiratory cassette.
Expiratory Cassette has a technical error
If this alarm message arises you must replace the expiratory cassette.
Expiratory Cassette Replacement Guide
  • gas supply
  • internal leakage
  • safety valve
  • O2 cell/O2 sensor
  • flow transducers
  • battery switch
  • alarm state
Cassandra Aarrestad
627 Reputation
4 Guides authored
Cal Poly, Team S7-G2, Paton Spring 2020 Member of Cal Poly, Team S7-G2, Paton Spring 2020
CPSU-PATON-S20S7G2
3 Members
60 Guides authored
5 Comments
Alarm: no O2 cell in use.
What's problem
Moshiur Rahman - Reply Share
air concentrator more than 25% during pre use check
sulaiman - Reply Share
Hii what's Internal error ...what can be the reason
Sandeep Raj - Reply Share
Why pressure transducer failed during preuse check ?
rajeswary subramaniam - Reply Share
Cassette error shown even after changing the cassette. Do you know why
Tony Dominic - Reply Share
Past 24 Hours: 2
Past 7 Days: 46
Past 30 Days: 106
All Time: 10,813