Medical Equipment
Neither iFixit nor this repair information is endorsed by or affiliated with the medical device manufacturer.
Student-Contributed Wiki
An awesome team of students from our education program made this wiki.
Newport HT70 Troubleshooting
Newport Medical Instruments Inc. Newport HT70 & Newport HT70 Plus Released March 4, 2010 Product Code: 73-CBK
Run off Battery Power
The battery of ventilator is dead sooner than expected.
Loose Power Outlet
Make sure to check the outlet with no OFF position. Losing power can be caused by the loose connecting wire to the outlet. Consider testing it in another outlet.
Broken Adapter
Check the adapter and power cable have any physical damage. Plug and unplug from the wall outlet and see the battery is charged. If it is broken or battery is not charging, consider changing either cable or adapter.
Replacement Battery
If the battery is resorting to back-up power after fully charged, it could be the battery problem. Inspect the battery for nay leak and make sure that it still has charged in it. If not, you might need to replace it with the new battery.
Circuit Problems
Ensure to plug in the correct transducing line to proximal line and exhalation valve and check all connections and humidifiers are functioning well.
Cleaning and Disinfection of Ventilator Circuit
If there is an obstruction such as fluid or dirt in the ventilator circuit, disassemble all pieces of the circuit and use unscented or non-antibacterial soap to wash those pieces. However, the exhalation valve and sensing wires should not be cleaned.
Air Intake Filter (Covidien HT460300)
Intake Filters should be checked weekly. Replace the filter when the surface area of the filter is no longer white. The Intake filter is in the right side of the ventilator behind the filter cover.
*****WARNING*****
The HT70 Ventilator should not be operated without a clean Air Intake Filter installed. Never re-use a dirty Air Intake Filter in reverse.
Failure of ‘Circuit check’
This is one issue that may arise from likely problems with the connection of the breathing circuit. Check to make sure that the breathing circuit is connected and does not have any leaks. If problem persists, user may want to seek out replacement of patient circuit and/or pump assembly.
System Error
The system displays that an error has occurred. This may be due to a number of calibration issues, review Event History Log for specific error. ''
Calibrations
HT70 calibrations should be performed for any repair, and for scheduled preventative maintenance, which is especially important to pay attention to during times of heavy usage (COVID-19). A number of system errors may occur if the calibrations of this machine are not maintained. The scheduled calibrations are as follows: Every 24 months or 15,000 hour service.
 Pump Leak Calibration Error
For system error pertaining to pump leak calibration, use the following steps:
1.     On the calibration screen, select ‘pump leak cal’. A message will appear asking you to ‘attach the rp50 to the patient outlet. Press the button when ready.’
2.     Make sure to connect the rp50 parabolic resistor to the patient outlet (Main flow outlet) (you may have to disconnect the breathing circuit to do this). DO NOT BLOCK OPEN END.
3.     Select ‘Rp50 Ready.’
4.     Wait for the ventilator to perform its calibration. The unit will increase speed slowly.
5.     If the calibration is passed, a message telling you that the calibration has passed should appear. Proceed by pressing the ‘Accept Cal’ button to save the calibrated value.
6.     If the test fails, press ‘cancel Cal’ and repeat calibration procedure.
7.     **IMPORTANT: Make sure to remove Rp50 from Main flow outlet and reconnect breathing circuit to Main flow outlet.
FiO2 Sensor Calibration
For System error pertaining to FiO2 Sensor, recalibrate by using the following steps:
1) Press the ‘Start O2 Cal’ button.
2) Press the ‘O2 Ambient Ready’ button.
      Note: During test the ventilator will display ‘Please Wait’.
         Once completed, the ventilator will state ‘O2 100% Ready’.
3)  Connect the HT70 Low Flow Oxygen Reservoir to the ventilator.
4) Connect the O2 gas source to your flow meter. (note: make sure
the flow meter adjustment knob is in the off position.)
5)  Turn on the oxygen source and turn the flow meter up to 10 L/min
to flush the reservoir.
6)  Press the ‘O2 100% Ready’ button.
      Note: During the test the ventilator will display ‘Please Wait’.
7) You will now need to adjust the flow meter in order to maintain a steady flow of O2 going through the system.
8)  Once the “Calibration Passed” shows on the message screen, you will need to adjust the flow meter down to zero and press ‘Accept Cal’ button on the main screen.
9) Make sure to disconnect all accessories from the air inlet port.
Motor Speed Calibration
If system error is pertaining to Motor speed or speed table, follow these instructions:
1) Disconnect the circuit from the output port
 2) On the calibration screen, select “Motor Speed Cal”.
3) The unit will slowly increase the speed of motor. Wait approximately 3 minutes for the ventilator to complete this calibration.
 *If the calibration is passed, the message will appear: ‘Calibration passed. Press the ‘Accept Cal’ button to save the cal value’.
4) Press the ‘Accept Cal’.
a) If the test fails, press ‘Cancel Cal’ and repeat the calibration procedure.
Non-Responsive Screen
Blank/Dark Screen
If the screen is displaying a dark or a blank image, then the system may have the Power Save ON setting causing the Utility Screen to be in sleep mode. Touch the screen or any of the buttons to “wake up” the screen.
Unresponsive Screen
If the screen does not respond to any touch, the screen might be locked by the Auto-Lock function. To unlock the screen, locate the lock icon on the screen. The lock icon is on the bottom right corner of the screen. Touch and hold the lock icon for 3 seconds, and the screen will unlock.
If the method above does not work, then the Ventilator is using the Basic Domain function. To resolve this issue, locate the Basic Domain Button. The Basic Domain button can be found on the bottom right corner of the screen. Once pressed, toggle the domain function Hospital (Hosp). Press the Accept button to update the changes.
Merek McGill
520 Reputation
5 Guides authored
Cal Poly, Team S7-G5, Paton Spring 2020 Member of Cal Poly, Team S7-G5, Paton Spring 2020
CPSU-PATON-S20S7G5
3 Members
72 Guides authored
5 Comments
I have a failure in 2 Newport HT70+, the flow sensor does not recognize me, but it does not give me the alarm of the disconnected sensor either, it simply shows me the same set value on the screen and starts searching to sense a value, it does not get it and it shows the value set again, and so it is every half hour, but the strange thing is that it doesn't give me any alarm... can anyone help me?
* Les install the software from scratch
* Calibrated and passed all calibrations
Ing. Jose Valdez (josevaldezmujica@gmail.com)
Ing. José C. Valdéz Mujíca - Reply Share
my problem is when I turn on the unit it just hangs on the HT70 and won't turn off first wait for the battery to drain. Is it possible that the board is the problem? alphiejimenez08@gmail.com
alphie jimenez - Reply Share
Hola Alphine Jimenez, la falla tuya es la Batería de reloj CR1225 que esta en la Placa que está agotada, reemplázala e inicia el ventilador, si no inicia entonces ya tendrías que correrle nuevamente el Software... saludos
Hola, tengo 2 ventiladores HT70, ambos tienen el mismo problema el cual es que se quedan congelado al iniciar en la pantalla de las 6 imágenes. Probé cambiando la SBC y si funciona. Entonces la falla está en la tarjet SBC. Hay forma de recuperar la SBC??? Tampoco me da la opción de entrar al modo de servicio para actualizar el software. Cualquier ayuda me la comparten a ing.elmerperez@gmail.com
Saludos y gracias....
Soporte Técnico Digimedical - Reply Share
Buen día mi problema es que el PEEP del ventilador no sube en ningún modo que queda en 2, alguien sabría cual es la manera de solucionarlo? Saludos y Gracias
Daniela Amador Perez - Reply Share
Past 24 Hours: 2
Past 7 Days: 12
Past 30 Days: 59
All Time: 3,211